PXI 6289 Accuracy Calculations
I'm trying to calculate the absolute accuracy of a 6289 reading using the formulas in the spec. In looking at the example on page 5 of the NI 628X specification the number at the bottom that is calculated for absolute accuracy is 980uV. Is that +/- 980 uv?
(4,102 Views)
Bicbehr,
After searching our website a little bit, I believe it is +/- 980 uV.
Here is some documentation that might help you.
How Do I Calculate Absolute Accuracy or System Accuracy?
Online Accuracy Calculator
Douglas Choisnet
(4,057 Views)
Thanks for the reply. I reviewed the same info on your website and think you are right.
(4,054 Views)
AND don't forget the page 2 of the spec if you read more than one channel!!
Since the settling time error(s) are not covered in the these caculations.
Greetings from Germany
Henrik
LV since v3.1
(4,028 Views)