Hi Howmean,
Your interpretation of the settling time is not wrong, but note that it's for 1 LSB.
National Instruments specifies the maximum rate at which an analog output device can change from one voltage to another. However, NI does not specify how big of a voltage change the device can make in-between updates. For example, if you output an array of voltages such as 2.3, 2.4, 2.5, 2.6… at an update rate of 2.86 MHz you will see the 200040 warning. However, because you’re only making small voltage changes of 0.1 V, you can ignore the error. If you output an array of voltages such as -10, 10, -10, 10… at an update rate of 2.86 MHz you cannot ignore the warning because of the large 20V swing in-between updates. The resulting signal will likely end up resembling -5.321, 4.465, -4.994, 5.012… rather than -10, 10, -10, 10… The device simply does not have enough time to change the voltage level before the next sample.
Therefore depending on the signal you are trying to output, you probably can ignore the warning.
Let me know if you have further questions.
Regards,
Boon Chen
05-31-2010 03:04 AM
Hi Boon Chen,
Thanks for your reply.
However, I have another question about the Digital Output using PXIe-6259 instead.
Please refer to my attached VI.
For the "Daqmx Timing (Sample Clock).vi", I have tried with both the "100kHzTimebase" and "10MHzRefClock" as the sample clock source.
I try to observe the changes in the digital output waveform by changing the "Sampling rate".
However, it looks like it does not affect the duration of the output waveform at all.
Since changing "Sampling rate" doesn't affect duration of the output waveform, I modified the "number of samples" instead.
By doing this I saw some changes.
May I know why is that so?
It is rather confusing because the relationship between Sampling rate and Number of samples in analog voltage generation case is always straight forward.
If I set the sampling rate to 100k S/s and number of samples to 10k. I expect to see an output waveform with duration of 100ms.
Thanks again.
Howmean
06-01-2010 02:22 AM
Hi Howmean,
Did you change the rate & number of samples for the DAQmx timing VI or did you change that of the sub Generate Dig WDT Data.vi?
Since you are only looking at line0 of the digital waveform, it's simply an alternating waveform between 0 and 1. If you did not change the settings for the sub Generate Dig WDT Data.vi, I expect that you will see the same waveform being output. Normally we use the same number of samples input for both the DAQmx Timing VI and the sub Generate Dig WDT Data.vi.
For a clearer picture, I would suggest that you refer to our shipping example named Write Dig Chan - Ext Clk.vi
Regards,
Boon Chen