   I'm using a USB-6218 device to read signals for studies about electrical power quality on “differential” input configuration and “on demand” mode. At first the device worked well until it has shown a problem. In the MAX program, appears a caution symbol ("!") over the device icon. When I use the test panel the screen doesn't show anything and an error-200474 is reported.

   If I reset the device through MAX a message that it was reseted is showed but the problem doesn't disappear. If I make a self-test, the device passes at the test in spite of the problem. When I try to calibrate USB-6218, the program stops working requiring me to disconnect the device from computer.

  Is this a software or hardware problem?

 

  Thanks.

 

0 Kudos
Message 1 of 5
(3,051 Views)

It sort of sounds like you blew the device after subjecting it to inputs greater than absolute maximum allowed.  How was the input referanced to ground?


"Should be" isn't "Is" -Jay
0 Kudos
Message 2 of 5
(3,048 Views)

Good morning,

Follow the links for the error that happens in your application, if not please send a "PrintScreen" error to be able to find a possible solution. Thank you.

Error -200474 With Digital Frequency Example VIs Modified for Continuous Acquisition

http://digital.ni.com/public.nsf/allkb/4E017925BDF2DC2D862572E6007C9594?OpenDocument

 

Why Do I Receive Error -200474 When Verifying an S Series Device In Calibration Executive?

http://digital.ni.com/public.nsf/allkb/F94FF852680CA9F78625739F007B3ECF?OpenDocument

 

Sincerely,

Mauro Vera
Applications Engineer
National Instruments
0 Kudos
Message 3 of 5
(3,039 Views)

Hi Guimaraes,


In the calibration process , you SHOULD disconnect the board to make a isolated test in MAX.
Try to realize this test and report here if you have any errors.

Paulo Jr. | Applications Engineering | Brazil
0 Kudos
Message 4 of 5
(3,038 Views)

I was reading a voltage signal from a voltage stabilizer (of course I conditioned the signal using a voltage divider). I decided to connect the AI GND to ground output of that device. I  have made a mistake?

 

Excuse for my English.

Thanks.

0 Kudos
Message 5 of 5