Question on using DAQCard-4350

I have a DAQCard-4350 (PCMCIA) and a PSH32-TC6 cable for taking thermocouple readings.  I want to use this card to take voltage readings.  The installation guide for the PSH32-TC6 (321637C-01) cable says on page one, “The pod provides connections to all NI PCMCIA-4350 DIO lines and to all but two analog input channels (CH)”.  How do I get access to the analog inputs?  

0 Kudos
Message 1 of 4
(3,075 Views)

J-Kane, 

 

Thank you for using NI forums! The two AI channels you do not have access to are CH0 which is dedicated to the cold-junction sensor and CH1 which is dedicated to auto-zeroing circuitry. It is implemented automatically, and thus limits your input channels to 14 instead of 16. With that cable, you cannot access those two channels.

 

Katie

Katie Collette
National Instruments
0 Kudos
Message 2 of 4
(3,059 Views)

J-Kane, 

 

If you would like to access those two additional channels, consider using a TBX-68

 

Katie

Katie Collette
National Instruments
0 Kudos
Message 3 of 4
(3,056 Views)

Thank you.  But I asked how I get access to the analog inputs.  I only need to input one channel but did not know how to input the voltage using the PSH32-TC6.  

 

Now I know.  By doing a little inspection of the cable and a voltmeter I have found that I have to take one of the mini-thermocouples and rempve the thermocouple wire and use copper wire.  Attach this copper wire to my voltage source and then I am reading voltages and not temperatures.

 