Multifunction DAQ
SCXI-1100 measurement accuracy
Dear Team,
I am trying to calculate measurement accuracy for SCXI-1100 (http://digital.ni.com/manuals.nsf/websearch/0790D3C92CF4B16E8625665E00712865).
This is not described in standard table, there is input / output stage accuracy (should be added ?). I can not achieve a accuracy which is a result of accuracy calculator. If you have any equatuions / examples please post.
Michal
(3,232 Views)
Accepted by mpankau
Hi Michal,
I am not sure that I understand your question. Do you try to calculate accuracy of SCXI-1100 but you do not have the specification data for the calculation, or you do not know how to calculate the accuracy?
If you need specification data you can check the SCXI 1100 data sheet. If you want to calculate your device accuracy, I can recommend you following KB: How Do I Calculate Absolute Accuracy or System Accuracy?
I hope this is the info you were looking for.
Barna D.
(3,180 Views)
Dear dobarni,
thank you for reply, in meantime I have found also this specification, which has more 'friendly' accuracy table than in manual,
and I did all caluculations which were needed.
Regards.
Michal
(3,163 Views)
