10-09-2013 02:17 PM
I am new to the NI world and have a project to measure TC's and RTD temperatures using a TBX-68T and NI 4351 DAQ.
I can not find a listing or diagram of how to connect the TC's or RTD's to the terminals.
can someone provide a diagram?
Thanks in advance.
10-10-2013 05:39 PM
Hello Dennistr,
You can use the manual of the 4351 as a reference: NI 4350/4351 User Manual. For thermocouples, on page 3-8, for RTD from page 3-16 to 3-18. There you can see the terminal name, then if you go to page B-3:5 to can find the relation between the terminal name and the label on the TBX terminal block.
Regards
Frank R.
10-11-2013 09:48 AM
Thank you for the information.