I'm trying to carry out three parallel voltage measures with a USB-6210 Card. Whatever the control software I'm using (vi under LabVIEW + DAQ, SignalExpress or Measurement and Automation Explorer), I get the same strange signal. In attachment please find an example from the control panel. As I'm a beginner with LabVIEW, I suspect some stupid mistake, but I checked my wiring and drivers without finding problems.
Thanks for your help
Best regards,
Saiko
Solved! Go to Solution.
02-22-2011 10:58 AM - edited 02-22-2011 10:59 AM
sorry, replied to wrong post
02-23-2011 04:50 AM
Hello Saiko,
Thank you for posting on National Instruments' forum.
This link (french post) will resolve your problem : here.
The problem was using a Differential mode (AI- AI+) to acquire an analog signal instead of RSE (AI GND).