Multifunction DAQ
is calibration possible for these devices?
I use Calexec3.1 for calibrating NI devices. I have a list of devices, which are not listed in the supported device types of the calexec. I need a confirmation from you that these devices can not be calibrated using Calexec. I would also like to know if NI provides a means to calibrate them if the parts are shipped to you. Following is the list of devices:
6534
8105
5402
8196
MXI-3
6533
8186
8185
7833R
8196
2503
5660
2596
5402
2593
2596
5660
2591
2530
8176
5411
2501
8331
6562
6534
(3,088 Views)
While I am not an expert at all of the devices you've listed, I can tell you that many of them are devices that do not require calibration.
Please see the attached spreadsheet.
Sal
(3,071 Views)
I have updated the remaining fields in Sal's xls file.  There is a complete list of calibration services NI offers at www.ni.com/calibrate.  A pdf if devices you can have calibrated at NI is here, and devices you can calibrate with Cal Exec is here.
I hope this helps,
Jesse O.
Applications Engineering
Jesse O. | National Instruments R&D