I'm using a NI 9219 in current mode to acquire several signals, and I need to calculate the measurement uncertainty due to the module.
In order to do so, I must know the resolution, the absolute accuracy and the sensitivity.
According to the manual (specifications start page 23) :
- Resolution is 24 bits
- Absolute accuracy is function of gain error and gain offset (For typical use : Absolute Accuracy = Gain error * measured value + gain offset * full range)
- Sensitivity is ???
How can I know the sensitivity of the module in current mode ?
Thank you & Regards
Noé
02-17-2016 10:09 AM - edited 02-17-2016 10:10 AM
Hi Noé,
If the sensitivity is not specified then it should be calculated dividing the range you use by the number of possible values.
As you use a 24 bits the result should be something like : (25 - (-25)) mA / 2^24 values.
I hope it helps even though I may be mistaking.