niSync Initialize

Solved!
Go to solution

I have a 9189 cDAQ chassis connected to my laptop and confirmed running in MAX. When I run niSync Initialize.vi I get error -1074118552: "A failure occured while communicating with the remote device. Ensure the remote device is accessible." I've tried using a direct IP address (sync://10.34.66.85/system) instead of he device name but the same error is returned. Also tried removing the insertion of .local. Any suggestions for how to correct this? The code below comes from the example attached to https://knowledge.ni.com/KnowledgeArticleDetails?id=kA03q000000x54zCAA&l=en-US.

 

Testr00_0-1629652657167.png

Testr00_1-1629652707516.png

Testr00_3-1629652847178.png

 

 

 

0 Kudos
Message 1 of 3
(1,795 Views)
Solution
Accepted by topic author Testr00

I'd recommend checking the firmware version of the cDAQ-9189. I believe 18.1.0 was the first version to support the NI-Sync API, but I'd recommend using the latest.

 

If you already have the latest firmware, could you list which versions of NI-DAQmx, NI-Sync and LabVIEW you are using?

 

Regards,

-Tyler

 

Related links

Does the NI-9185/9189 Support PTP/IEEE 1588v2?

 

Message 2 of 3
(1,749 Views)

Works after updating 9189 firmware. Thanks!

0 Kudos
Message 3 of 3
(1,732 Views)