Hi Shawn,
Verify that you are seeing this behavior in Test Panels. It seems you are using pins 68 and 67 based on your description, which is a reference single ended (RSE) configuration. Ensure you are making this distinction in your setup. I would also recommend trying other AI channels. Hopefully this helps!
09-26-2010 06:36 PM
09-27-2010 11:55 AM - last edited on 05-02-2024 08:58 AM by Content Cleaner
Whether or not you need to tie ground your AIGnd will depend on whether the transducer is floating or grounded- see the wiring guide. All of the ground channels on the board are tied together so whichever way is easiest for you. Pages 60 and 61 would also be helpful in determining the appropriate grounding configuration and whether or not your DAQ board needs to tie to ground.
09-28-2010 09:37 AM
09-28-2010 11:15 PM
10-01-2010 07:08 PM - last edited on 05-02-2024 09:00 AM by Content Cleaner
There are a couple things to address. 1) Reading an open channel will usually result in an erratic signal as nothing is connected. 2) Using a strain gauge will require excitation that cannot be provided by the PCI-6013e. You will need to an external excitation source. 3) The cable you are using is a 68 pin to 50 pin cable, which means you cannot use the pinouts from the specifications but rather you should reference the pinouts for this cable. Typically the SH68-68-EP (2m) - 184749-02 is used with E, M, and X series devices. This should help clarify the behavior you were experiencing.