Multifunction DAQ
"preload" NI USB-6251 Input Channels
I have a number of different measurements happening on the 6251 device. The analogue input voltage varies from 0.2V to 4V in the worst case.
When I configure the instrument to take the low voltage measurement just before the low voltage measurement, I can see the reading frequency on the low signal. To get a better result, I use a second channel to measure the same low signal.
At the moment it looks like this
Ch1: 4V
Ch2: 0.2V
Ch3: 0.2V
In this case the internal voltage is already close to the input voltage I expect on Ch3.
Is there a way to solve this in the code? Some set the internal voltage to value I expect at the input before making the measurement?
(709 Views)
Accepted by me1ody
Have you included the recommended bias resistors (generally 10k-100k) between your signal terminals and AI Gnd?   See this article for details (plus a lot more good info).
-Kevin P
(691 Views)
How did you measure the "internal voltage"? Did you connect it to a 0V signal? Or did you leave the channel open? Please noted that an unconnected channel does not return 0V in NI-DAQ. See Incorrect Readings on Unconnected or Open Channels of DAQ Device
Control Lead | Intelline Inc
(672 Views)
I didn't measure it, sorry. I just thought that's the case.
(653 Views)
