PXI
Cascading Analog Outputs
I found a link establishes that you can cascade the analog output channels of particular devices (e.g. PXIe-4322)...
https://knowledge.ni.com/KnowledgeArticleDetails?id=kA00Z0000019RIsSAM&l=en-US
(3,708 Views)
You can also review the User Manual of the device, such as the PXIe-4322's 2nd appendix, and find references to these configurations if they are supported.
(3,657 Views)
So I have no idea what happened to my original post, but I intended for it to have a question 🙂 (Might have been holiday fever!). The link states the following:
  • To get the stacked voltage on the Load, you will need to set each channel being stacked to a portion of that total voltage. For example, in the figure the load needs a total of 32 volts, so channel 0 and channel 1 needs to be set to output 16 volts each.
That seems straight forward for a DC voltage output. What I'm curious about is whether or not you could generate a sine wave with that same stacked voltage configuration? Could you simply take an array of points that would represent the sine wave amplitudes, divide that by 2, and send the result to each channel and expect a clean sine wave?
(2,300 Views)
Yes you can.  At least, I am doing it as a differential output with a PXIe-6738.  Just make sure both channels are in the same task and they will be updated together.
GCentral
(2,295 Views)
So just to be more clear, the PXIe-4322 has a +/- 16V per channel. The link seems to say that if you wanted a 32V output, you can "stack" two channels together (see diagram in link), specify an analog output of 16V on each channel, and effectively get 32V. The stacking is allowed due to channel to channel isolation. Can you do that with the PXIe-6738?
(2,291 Views)
