Status Error Code -50400, PXI 4461

Solved!
Go to solution

Hi all,

 

I am using NI PXI-4461 for testing. But suddenly today our labview program stops working now with a status code -50400, which says 'The transfer did not complete within the timeout period or within the specified number or retries.' I could not find the meaning of this status code. I do need your help.

 

Any thoughts will be appreciated.

 

Regards,

 

Steve

Download All
0 Kudos
Message 1 of 4
(1,807 Views)
Solution
Accepted by topic author xzzgameboy

Since the error -50400 occurred during self-test, I would consider it as the instrument did not respond back within the required timeout.

 

Suggested troubleshooting options,

  • Reset, Self-test, self-calibrate
  • Power cycle the PXI system
  • Move the 4461 to a different slot

 

If the module is still not functional after these, it could be just a case of a damaged/bricked instrument that needs repair by NI.

Santhosh
Soliton Technologies

New to the forum? Please read community guidelines and how to ask smart questions

Only two ways to appreciate someone who spent their free time to reply/answer your question - give them Kudos or mark their reply as the answer/solution.

Finding it hard to source NI hardware? Try NI Trading Post
0 Kudos
Message 2 of 4
(1,767 Views)

Thanks!!!! 

 

I did reboot the system and the second time somehow this error went away.

 

Not sure what caused this issue but now the system is working properly.

 

Thanks again!

 

 

0 Kudos
Message 3 of 4
(1,765 Views)

Hi santo_13, today I had the same problem but in my case there was an error also for resetting (pict attached). I repeated the same steps in another slot and with a new board.
Do you have other suggestions?

 

Thank you