10-11-2012 01:32 PM
What signal source can you use with the 1300 (ie. no bias resistors)? I am having an issue with ground referenced signal on the 1300.
10-12-2012 09:29 AM
BME_Genius,
What SCXI module are you using along with the 1300 terminal block? Also, what type of signal are you trying to measure?
Regards,
10-12-2012 09:54 AM
I am using the 1100 module. The signal is a 0-7.5 Vdc from a CO2 analyzer. The output impedance is 20kohm.
10-15-2012 03:01 PM
Have you verified the jumper settings in the 1100 module correspond the the grounding type of the signal source? A description of the jumpers is available in the 1100 user manual:
http://www.ni.com/pdf/manuals/320637c.pdf
Regards,
10-16-2012 07:35 AM
Yes. Jumper W1 is in position 1-2. However, it cannot be a module problem because it only occurs when that signal is attached to the 1300. This was never a problem before with this signal on 1300 and started to occur intermittently and then would not go away. The problem signal still is and always was ground referenced.
10-17-2012 04:51 PM
Interesting...
It seems that a ground-referenced signal shouldn't need any special wiring considerations...
From the manual:
"If you are using a ground-referenced thermocouple, connect the
thermocouple positive and negative leads to CH+ and CH– respectively.
Leave jumper W1 on the SCXI-1100 in the factory-default (parking)
position"
This is how you have if wired, but you're still having his issue, correct?
Also, does this behavior hold regardless of input channel?
Regards,
10-18-2012 06:47 AM
Yes, this is how I have it wired and always has been wired on the 1300. No matter what channel I use on the 1300 I see the same problem. For some reason, it works fine on the 1303. That gave me the idea to tie the negative terminal to ground via a 10kohm resistor. This fixes the problem on the 1300...but WHY being it is ground referenced????