How can I filter thermocouple noise using a TC-2095 and SCXI-1102c?

Solved!
Go to solution

Hello all,

 

I currently have four TC-2095s connected to SCXI-1102c cards inside a SCXI-1001 chassis. These thermocouples are floating and I have been getting a fair amount of noise (3-phase 400 Hz and even 60 Hz noise) when we power on some equipment such as blowers and fans. Using this current set up, how can I filter the noise? Or do would I need to set up a new hardware configuration (i.e. SCXI-1102 cards)?

0 Kudos
Message 1 of 3
(5,798 Views)
Solution
Accepted by topic author rlyles

Hi RALoo, 

 

The 1102C has a lowpass filter with a cutoff at 10kHz, so you will not be able to filter in hardware with your current setup.  

Your first option would be to use a software filter.  LabVIEW has several functions which accomplish this. More information on this can be found here: http://zone.ni.com/reference/en-XX/help/371361J-01/lvanls/filter_vis/

 

The second option would be to change to SCXI 1102 cards as you suggested.  The lowpass filter of the 1102 has a selectable cutoff frequency at 2 Hz, which will stop the 400 and 60Hz noise from affecting your readings.

Patrick W.
Applications Engineer
National Instruments
Message 2 of 3
(5,784 Views)

Thank you Patrick! I suspected that was the case.