08-04-2020 04:55 AM
Hello everyone,
Do I need any external bridge completion module, like NI-9949, for reading a non-amplified strain gauge output using NI-9235 module, or the NI-9235 module includes built-in bridge itself?
Thanks in advance
08-04-2020 08:10 AM - edited 08-04-2020 08:12 AM
The NI 9235 provides connections for eight 120 Ω quarter-bridge input channels.
so ... as long as you have a quater-bridge , no problem, connect it as shown in the fantastic ( 😉 ) manual
What type of sensor do you want to connect?
08-04-2020 08:16 AM
Hi Henrik,
Thanks for your response. Yes, that is it. I'm gonna connect a 120 ohms strain gauge in quarter-bridge config.