09-10-2014 07:48 PM
I am using a myRIO to measure strain with a Wheatstone bridge. I used the wiring scheme shown in Fig 4 at http://www.ni.com/example/31417/en/, except wired the Vex to C/AO1 instead of the constant 5V. When I run my VI, I get a measured Vex of only 0.2V for anything from 1-10V. If I disconnect my Wheatstone bridge and just connect straight from AO to AI, it measures the corresponding voltage (using the same VI). I am wondering if there is some current limit on the MSP on myRIO that is cutting my voltage? Or is there something else going on?
Thanks!
09-11-2014 10:38 AM
Hi Zillion,
The myRIO has a 3mA current limit on the Analog Outputs. These are not meant to supply power. This can be seen on page 22 of the myRIO specifications manual here.
Hope this helps!
09-12-2014 10:03 AM
Using external power solves the problem. Thanks!