Can a DMM be setup to get data between start and stop triggers?

I need to get data with a PXI-4071 between the rising and falling edges of a trigger signal. Can that be done?

 

George

0 Kudos
Message 1 of 4
(6,687 Views)

Hi George,

       First of all, what is your trigger?  It is possible to sample from your 4071 as long as your trigger signals are within the reading rate specifications of the 4071.  The max reading rate is listed on the specs, 10kS/s for 4.5 digits of resolution.   Otherwise, I would highly recommend checking out the following documentation on DMM triggering:

 

PXI-4071-DC Voltage DMM/Switch Handshaking Scan Rates

If you're using a switch: Multi-Module Switching with National Instruments Switches

 

Let me know if you still have questions!

 

0 Kudos
Message 2 of 4
(6,671 Views)
My trigger will be a 0 to 3V signal that stays high anywhere from 100 msec to 3 sec. Basically I want to use my signal as a start / stop trigger. I don't see any way to implement both a start and stop trigger.
0 Kudos
Message 3 of 4
(6,649 Views)

I assume you are referring to a digital trigger being imported to the instrument via one of the PXI trigger lines or the front panel connector?

 

None of NI's digital multimeters support a stop trigger. You can acquire a finite number of measurements after receiving a trigger. Perhaps you can combine this with other hardware to know when to stop acquiring or what measurements to throw away.

 

Marcos Kirsch
Chief Software Engineer
NI Driver Software
0 Kudos
Message 4 of 4