Struggling with setting up a RS-485 connection from a mass flow controller to NI cDAQ
I am an absolute beginner starting from the ground up trying to connect the RS-485 port from a Sierra Instruments mass flow controller (MFC) to the NI cDAQ.
The intended pathway that someone before me laid out for the connection is:
1) Cable from MFC COM1 to the 4-Port Ethernet Interface for RS485/RS422 (DB-9 to DB-9)
2) Ethernet cable from 4-Port Ethernet Interface for RS485/RS422 to NI cDAQ
I am struggling to understand exactly how the pins should be lined up for the 1st part of the interface. The pinout for MFC COM1 is given below:
The pinout for DB-9 male shown in NI's Serial Quick Reference Guide is shown below:
I am not sure which pins at the MFC side should connect to which pins at the ENET-to-485 side. The obvious one is ground-to-ground (Pin 7 to Pin 1). Beyond that I'm stumped. Trying to read up references online has just served to confuse me even more at this point...
Any help would be much appreciated!
(5,385 Views)
It looks like the vendor stuffed a lot of signals into a couple of DB9 connectors besides the RS485 signals. Do you have the manual that you can attach?
(5,356 Views)
I think I figured out the pin connections I need to make from one end to the other if I choose to go this 9-pin to 9-pin route. It appears the MFCs are 2-wire RS-485 while the receiving end (NI terminal) is setup for 4-wire RS-485. So I will need to wire the TxD+ & RxD+ at the NI terminal to RS-485-A pin of MFC COM1, and wire the TxD- & RxD- at the NI terminal to RS-485-B pin of MFC COM1. The other signals (ID-0, etc.) are optional for the MFC & the power will be supplied to the MFCs through a separate connection. So that takes care of all the necessary connections.
I've also identified another pathway into the computer using an USB to RS-485  converter. I think at the end of the day this will be an easier setup to follow. Thanks for the reply & I will see where this takes me.
(5,318 Views)
Accepted by topic author Mango17
To follow-up: I purchased a B&B Electronics ULinx USB to 9-pin RS-485 converter as well as a cable from Sierra Instruments. I was able to confirm communication between the cDAQ and the MFC using Sierra's own HMI provided as well as through NI-MAX.
(5,183 Views)