Struggling with setting up a RS-485 connection from a mass flow controller to NI cDAQ

Solved!
Go to solution

I am an absolute beginner starting from the ground up trying to connect the RS-485 port from a Sierra Instruments mass flow controller (MFC) to the NI cDAQ.

 

The intended pathway that someone before me laid out for the connection is:

 

1) Cable from MFC COM1 to the 4-Port Ethernet Interface for RS485/RS422 (DB-9 to DB-9)

2) Ethernet cable from 4-Port Ethernet Interface for RS485/RS422 to NI cDAQ

 

I am struggling to understand exactly how the pins should be lined up for the 1st part of the interface. The pinout for MFC COM1 is given below:

485 pinout.PNG

 

The pinout for DB-9 male shown in NI's Serial Quick Reference Guide is shown below:

485 pinout.PNG

 

I am not sure which pins at the MFC side should connect to which pins at the ENET-to-485 side. The obvious one is ground-to-ground (Pin 7 to Pin 1). Beyond that I'm stumped. Trying to read up references online has just served to confuse me even more at this point...

 

Any help would be much appreciated!

 

0 Kudos
Message 1 of 4
(5,218 Views)
It looks like the vendor stuffed a lot of signals into a couple of DB9 connectors besides the RS485 signals. Do you have the manual that you can attach?
0 Kudos
Message 2 of 4
(5,189 Views)

I think I figured out the pin connections I need to make from one end to the other if I choose to go this 9-pin to 9-pin route. It appears the MFCs are 2-wire RS-485 while the receiving end (NI terminal) is setup for 4-wire RS-485. So I will need to wire the TxD+ & RxD+ at the NI terminal to RS-485-A pin of MFC COM1, and wire the TxD- & RxD- at the NI terminal to RS-485-B pin of MFC COM1. The other signals (ID-0, etc.) are optional for the MFC & the power will be supplied to the MFCs through a separate connection. So that takes care of all the necessary connections.

 

I've also identified another pathway into the computer using an USB to RS-485  converter. I think at the end of the day this will be an easier setup to follow. Thanks for the reply & I will see where this takes me.

0 Kudos
Message 3 of 4
(5,151 Views)
Solution
Accepted by topic author Mango17

To follow-up: I purchased a B&B Electronics ULinx USB to 9-pin RS-485 converter as well as a cable from Sierra Instruments. I was able to confirm communication between the cDAQ and the MFC using Sierra's own HMI provided as well as through NI-MAX.

0 Kudos