Bug in Driver Keithely 24xx
Hi
there seems to be a small bug in the driver for the Keithley 24xx. I'm not sure how to find out whom to mail this too (is there a contact for bug-reports listed somewhere?), so I 'm posting it here.
The bug is in Keithley 24xx.lvlib: Configure Settling Delay.vi
The string "%.;Sour:Del\s%g;" should read "%.;:Sour:Del\s%g;" that is with an extra ":" after the first ";".
That's it already 😉
Arun
(3,202 Views)
Thanks for letting me know, I’m going to see if I can get a hold of this instrument and reproduce it. What instrument are you using and what behavior do you see/expect to see?
Matt
Applications Engineer
(3,174 Views)
I'm using a Keithley 2400. The expected behaviour of that VI is to set the settling time between measurements. I get an error from the instrument (113). Adding the ":" fixes this problem and as far as I know the protocoll also wants a ":" over there...
HTH
Arun
(3,136 Views)
Thanks for letting me know the device you are using; unfortunately there is not an instrument available to me. I’ll check with R&D and see if there is an instrument there that I could test with. Can you post a screen shot or text of the error 113 that you get back from the instrument?
Matt
(3,116 Views)
Matt,
I recently worked on a program to use the Keithley 2400 and had problems with the driver also.
From the manual:
Eliminating common SCPI errors
There are three SCPI errors that occur more often than any others:
• -113, “Undefined header”
• -410, “Query INTERRUPTED”
• -420, “Query UNTERMINATED”
The following paragraphs discuss the most likely causes for these errors and methods for
avoiding them.
This error indicates that the command you sent to the instrument did not contain a recognizable
command name. The most likely causes for this error are:
Missing space between the command and its parameter. There must be one or more
spaces (blanks) between the command and its parameter. For example:
:SENS:VOLT:DC:RANGl00 Incorrect (no space between command and
parameter)
:SENS:VOLT:DC:RANG 100 Correct
Improper short or long form. Check the command list in Section 18 of this manual.
Blanks (spaces) within the command name. For example:
:SYST :ERR? Incorrect (space between :SYST and :ERR?)
:SYST:ERR? Correct
I saw all three of those errors repeatedly.  Eventually I gave up and started to write a new driver just using the SCPI commands and VISA. Little syntax errors seemed to be a part of the problem.  Unfortunately the project was cancelled just before I got things working, so I do not have any improved drivers to contribute.
Lynn
(3,112 Views)
Sorry, no time to upload a screenshot. The bug seems to be a typo according to the SCPI commands in the Keithley manual (they all start with a ":"). You can just compare that command in the driver with the one stated in the manual "Series 2400 Source Meter Quick Start Guide" which I believe is available for download from Keithley. BTW: is there a bug-tracker for each driver or a Changelog, VC-repository or something else? An email of a contact person for bugs would be great, too...
ARUN
(3,101 Views)
Thanks for the information. I’ll pass this information along to R&D who will look more closely at this instrument driver and make changes accordingly. I don’t know of an external changelog or VC repository that is available outside of R&D. If you think you have found a bug, in the future you go to ni.com/ask and create a service request in order to get into contact with an engineer.
Matt
(3,081 Views)