Bug in Driver Keithely 24xx

Hi

there seems to be a small bug in the driver for the Keithley 24xx. I'm not sure how to find out whom to mail this too (is there a contact for bug-reports listed somewhere?), so I 'm posting it here.


The bug is in Keithley 24xx.lvlib: Configure Settling Delay.vi

The string "%.;Sour:Del\s%g;" should read "%.;:Sour:Del\s%g;" that is with an extra ":" after the first ";".

That's it already 😉

Arun

0 Kudos
Message 1 of 7
(3,201 Views)

Hi,

Thanks for letting me know, I’m going to see if I can get a hold of this instrument and reproduce it. What instrument are you using and what behavior do you see/expect to see?

Matt
Applications Engineer
National Instruments
0 Kudos
Message 2 of 7
(3,173 Views)

I'm using a Keithley 2400. The expected behaviour of that VI is to set the settling time between measurements. I get an error from the instrument (113). Adding the ":" fixes this problem and as far as I know the protocoll also wants a ":" over there...

 

HTH

 

Arun

0 Kudos
Message 3 of 7
(3,135 Views)

Hi,

Thanks for letting me know the device you are using; unfortunately there is not an instrument available to me. I’ll check with R&D and see if there is an instrument there that I could test with. Can you post a screen shot or text of the error 113 that you get back from the instrument?  

Matt
Applications Engineer
National Instruments
0 Kudos
Message 4 of 7
(3,115 Views)

Matt,

 

I recently worked on a program to use the Keithley 2400 and had problems with the driver also.

 

From the manual:

Eliminating common SCPI errors

There are three SCPI errors that occur more often than any others:

• -113, “Undefined header”

• -410, “Query INTERRUPTED”

• -420, “Query UNTERMINATED”

The following paragraphs discuss the most likely causes for these errors and methods for

avoiding them.

-113, “Undefined header”

This error indicates that the command you sent to the instrument did not contain a recognizable

command name. The most likely causes for this error are:

Missing space between the command and its parameter. There must be one or more

spaces (blanks) between the command and its parameter. For example:

:SENS:VOLT:DC:RANGl00 Incorrect (no space between command and

parameter)

:SENS:VOLT:DC:RANG 100 Correct

Improper short or long form. Check the command list in Section 18 of this manual.

Blanks (spaces) within the command name. For example:

:SYST :ERR? Incorrect (space between :SYST and :ERR?)

:SYST:ERR? Correct

 

 

I saw all three of those errors repeatedly.  Eventually I gave up and started to write a new driver just using the SCPI commands and VISA. Little syntax errors seemed to be a part of the problem.  Unfortunately the project was cancelled just before I got things working, so I do not have any improved drivers to contribute.

 

Lynn 

0 Kudos
Message 5 of 7
(3,111 Views)

Sorry, no time to upload a screenshot. The bug seems to be a typo according to the SCPI commands in the Keithley manual (they all start with a ":"). You can just compare that command in the driver with the one stated in the manual "Series 2400 Source Meter Quick Start Guide" which I believe is available for download from Keithley. BTW: is there a bug-tracker for each driver or a Changelog, VC-repository or something else? An email of a contact person for bugs would be great, too...

 

ARUN

 

 

0 Kudos
Message 6 of 7
(3,100 Views)

Hi,

Thanks for the information. I’ll pass this information along to R&D who will look more closely at this instrument driver and make changes accordingly. I don’t know of an external changelog or VC repository that is available outside of R&D. If you think you have found a bug, in the future you go to ni.com/ask and create a service request in order to get into contact with an engineer.

Matt
Applications Engineer
National Instruments
0 Kudos
Message 7 of 7