Hello,
I get the following error message with the Keithley DAQ 6510.
The voltage measurement works without any problems, whereas I get the attached error when measuring the temperature. The VIs are provided drivers for the Keithley model. I have tried troubleshooting with NI IO Trace, but have not been able to figure it out. Does anyone have a tip for me?
Thank you very much.
06-11-2021 09:34 AM
I see "Fehler", so I assume you are on a computer with German settings.
Often instruments drivers are not written for a system with a comma as the decimal point character.
Change your LabVIEW settings to NOT use the localized decimal point and try again.
Regards, Jens
06-13-2021 04:42 AM
Thanks for the quick reply. Unfortunately, it did not work. Is it possible that a LabVIew version with a different language would work?
Best regards
Özgür D.
06-13-2021 10:22 AM - edited 06-13-2021 10:24 AM
Hi Özgür,
can you debug the Keithley VIs to find the offending string, which triggers the error -151?
Once you have found the wrong command you can compare it with the description in the manual and finally correct the device driver VIs to avoid that error…
(Remark: As indicated by Jens' suggestion device drivers often fail when the computer does not use the English numeric formats. I also observed such problems regularly. And I learned to revisit device drivers as they most often don't do what I really needed…)