06-10-2013 07:56 AM
I am using an NI 9239 to acquire analogue input data. Using the DAQ Assistant I set acquisition mode N Samples, 8k samples to read and 8k acquisition rate. I wire the data output from the DAQ Assistant directly to a waveform indicator.
I expect the dt for the waveform to be 0.000125. I get 0.000120, which is an acquisition rate of 8333.333Hz!
What am I doing wrong?
Thank you.
Solved! Go to Solution.
06-10-2013 08:12 AM
You have to read the specifications of the 9239 to answer your question. Focus on page 18/19.
Additional hint:
You specify the MINIMUM sample rate for the acquisition in the DAQ Assistant.
hope this helps,
Norbert
06-10-2013 08:25 AM
thanks!