I am generating sine wave using "simulate signal" block and sending it to DAQ Assistant to excite a shaker. The settings of Simulate signal and DAQ assistant are shown and the VI is also attached. The frequency being sent to DAQ is same. BUt when I am measuring the output frequency of shakr it is always coming higher. I am using BNC NI 9260 card and the shaker is a pssive device.
Can someone please guide where the problem is?
Thanks and regards
Solved! Go to Solution.
01-31-2017 01:26 PM
I am generating sine wave using "simulate signal" block and sending it to DAQ Assistant to excite a shaker. The settings of Simulate signal and DAQ assistant are shown and the VI is also attached. The frequency being sent to DAQ is same. BUt when I am measuring the output frequency of shakr it is always coming higher. I am using BNC NI 9260 card and the shaker is a pssive device.
Can someone please guide where the problem is?
Thanks and regards
01-31-2017 01:46 PM
How do you know that the shaker is at a higher frequency?
01-31-2017 01:48 PM
I am measuring it using a laser displacemnet sensor and getting voltage through Labview in another computer
01-31-2017 02:27 PM
The frequency (samples/second) of the output signal have to match the speed of the hardware device output update rate. Check to see if you can verify the two match-up.
Ben
02-01-2017 03:20 AM
How much off is your measured frequency (in %, ppm or Hz) and is the error within the accuracy of your interferometer and/or DAQ board?
... and why is your DAQ Assistant configured to generate 1001 sample blocks?
02-01-2017 03:51 AM
Looking at the datasheet you will find:
So, if you create samples for a different samplerate, not supported by the device, the resulting frequency will differ 🙂
Your example uses 1k SPS , and I don't know how the driver will coerce this.
Look for the examples, I bet there is a vi 😉
Finally, you will always find some differences between the frequencies, because the two units you use (one for generating, another to measure) don't share the same clock.
02-01-2017 09:51 AM
My frequency is approx 1.61 times higher and the error is almost negligible. i just kept one sample extra to write to start the new wave with 1st point of new cycle however i m not sure about this.
02-01-2017 10:10 AM
As Henrik Volkers pointed out earlier your module can only sample at specific rates and as stated on Henrik's datasheet screenshot, the lowest rate is 51200 / 31 ~= 1.652 kS/s. That is why it is running ~1.65 times too fast.
To generate a theoretically perfect 1 Hz you either need to generate (and repeat) 31 periods of signal using 51200 samples or probably easier, 1 period of 51200 samples but change your rate (on both your Create Signal and your DAQ Assistent) to 51.2 kS/s
02-01-2017 10:15 AM - edited 02-01-2017 10:22 AM
@MAnees wrote:
My frequency is approx 1.61 times higher and the error is almost negligible. i just kept one sample extra to write to start the new wave with 1st point of new cycle however i m not sure about this.
According to the spec the lowers possible samplefrequency is 1.652 kHz , your output is a distorted sinewave with 1001 points...(assumed, I can't open your vi ) ... so I would expect 1.6503 kHz ... almost fits 😄
How do you measure the frequency?
Try it with a samplerate of 1652 Hz and generate 1652 sample 😉
(for 1Hz, if you need to generate 0.8Hz you need to recalc the number of samples 😉 )
EDIT: LocalDSP... beat me with 5min 😉
MAness: You should really learn some basics on digital signal processing. The hint with a higher samplerate is a good one!