LabVIEW

cancel
Showing results for 
Search instead for 
Did you mean: 

Trigger channels cDAQ 9239

Solved!
Go to solution

Hi all,

I am curious to know if it would be somehow possible to accomplish the following task:

 

I have a cDAQ-9174 module with a NI9239 voltage reader. I am wiring 2 of the 4 channel available in the following way: the first channel is getting input from the TTL signal from a laser source. the second channel is recording the current from a generic sample (ex. a photodiode (PD)).

My question: is it somehow possible to trigger the second channel (PD, when to acquire the signal) by getting the "trigger" signal from the first channel (TTL signal)?

 

So I can acquire only when the laser pulse is actively hitting the PD or a generic sample.

 

Thanks

E

0 Kudos
Message 1 of 8
(836 Views)
Solution
Accepted by topic author EdoardoA

Unfortunately, NI-9239 does not support analog triggers and cDAQ-9174 does not have any PFI ports for triggering. You would need a digital module. Reference: Setting Up a Digital Trigger in a CompactDAQ Chassis

-------------------------------------------------------
Control Lead | Intelline Inc
0 Kudos
Message 2 of 8
(827 Views)

Thanks for the quick reply!

Can you recommend me a digital module?

 

E

0 Kudos
Message 3 of 8
(821 Views)

I believe NI-9401 is the cheapest option for a TTL signal.

You can browse Digital I/O product page.

-------------------------------------------------------
Control Lead | Intelline Inc
0 Kudos
Message 4 of 8
(791 Views)

Is there any option for a module with BNC connections?

0 Kudos
Message 5 of 8
(738 Views)

If you spend some time exploring Digital I/O product page, you should be able to notice which module has BNC connectors.

Among them, NI-9402 has BNC connectors and supports LVTTL. LVTTL is compatible with TTL for input. According to NI-9402 Specifications

ZYOng_0-1701785955651.png

 

-------------------------------------------------------
Control Lead | Intelline Inc
0 Kudos
Message 6 of 8
(729 Views)

Thanks for help, I did check and I found that one too, I want to have a confirmation from you:) by the way, would it be possible instead of triggering the NI9239 with the NI9402 just use the NI9402 and with 2 inputs?

 

For instance, trigger in channel 1 and analog signal on channel 2 of the same module (NI9402)? Measuring the current from photodiode on channel 2.

 

Thanks,

E

0 Kudos
Message 7 of 8
(696 Views)

@EdoardoA wrote:

Thanks for help, I did check and I found that one too, I want to have a confirmation from you:) by the way, would it be possible instead of triggering the NI9239 with the NI9402 just use the NI9402 and with 2 inputs?

 

For instance, trigger in channel 1 and analog signal on channel 2 of the same module (NI9402)? Measuring the current from photodiode on channel 2.

 

Thanks,

E


NI-9402 is a digital I/O module.

-------------------------------------------------------
Control Lead | Intelline Inc
0 Kudos
Message 8 of 8
(666 Views)