12-04-2023 12:35 PM
Hi all,
I am curious to know if it would be somehow possible to accomplish the following task:
I have a cDAQ-9174 module with a NI9239 voltage reader. I am wiring 2 of the 4 channel available in the following way: the first channel is getting input from the TTL signal from a laser source. the second channel is recording the current from a generic sample (ex. a photodiode (PD)).
My question: is it somehow possible to trigger the second channel (PD, when to acquire the signal) by getting the "trigger" signal from the first channel (TTL signal)?
So I can acquire only when the laser pulse is actively hitting the PD or a generic sample.
Thanks
E
Solved! Go to Solution.
12-04-2023 01:11 PM
Unfortunately, NI-9239 does not support analog triggers and cDAQ-9174 does not have any PFI ports for triggering. You would need a digital module. Reference: Setting Up a Digital Trigger in a CompactDAQ Chassis
12-04-2023 01:21 PM
Thanks for the quick reply!
Can you recommend me a digital module?
E
12-04-2023 02:16 PM
I believe NI-9401 is the cheapest option for a TTL signal.
You can browse Digital I/O product page.
12-05-2023 05:35 AM
Is there any option for a module with BNC connections?
12-05-2023 08:19 AM
If you spend some time exploring Digital I/O product page, you should be able to notice which module has BNC connectors.
Among them, NI-9402 has BNC connectors and supports LVTTL. LVTTL is compatible with TTL for input. According to NI-9402 Specifications
12-07-2023 03:34 AM
Thanks for help, I did check and I found that one too, I want to have a confirmation from you:) by the way, would it be possible instead of triggering the NI9239 with the NI9402 just use the NI9402 and with 2 inputs?
For instance, trigger in channel 1 and analog signal on channel 2 of the same module (NI9402)? Measuring the current from photodiode on channel 2.
Thanks,
E
12-11-2023 01:58 PM
@EdoardoA wrote:
Thanks for help, I did check and I found that one too, I want to have a confirmation from you:) by the way, would it be possible instead of triggering the NI9239 with the NI9402 just use the NI9402 and with 2 inputs?
For instance, trigger in channel 1 and analog signal on channel 2 of the same module (NI9402)? Measuring the current from photodiode on channel 2.
Thanks,
E
NI-9402 is a digital I/O module.