02-18-2016 09:42 PM
1. What is the maximum voltage of analog signal? The data sheet says 10V range. What exactly does range mean? I know it seems like a pretty silly question but I would really appreciate your help.
2. After establishing signal, how do I go about calibrating this?
Solved! Go to Solution.
02-18-2016 10:02 PM
Ok never mind guys, I found out what range means. My question is, what is the minimum voltage required for it to run?
Also, how do I go about calibrating it?
02-18-2016 10:45 PM - edited 02-18-2016 10:59 PM
6009 is 14 bit card with max range of ±10 v for Single-ended and ± 20 V range for Differential it means if you input voltage out of this range it will be show max saturated data(±10 or ± 20 V) also it is dangerous to overflow max voltage and you have to care about it
this card it standard card with 5 mv standard system noise (0.5mv for Differential in 1 v max setting)
02-18-2016 10:49 PM
02-19-2016 12:25 AM
Hi hatef,
Thank you for taking the time to reply. Could you clarify how you came to the answer 0.0012?
02-19-2016 12:32 AM
Sorry Hatef, I understood how you got the asnwer.
How do you go about calibrating it?
02-19-2016 12:45 AM
02-19-2016 01:00 AM
Again, my apologies. I have to calibrate an unltrasonic sensor, not the USB 6009.
02-19-2016 01:08 AM