I have multiple cRIO-9081. Each cRIO-9081 is running NI-DAQmx 15.5. On four of the cRIO-9018 we have attached cDAQ-9178s with 8 modules. Everything just worked. On the fifth cRIO-9018 we attempted to connect a cDAQ-9174. The device showed up in MAX and Windows Device Manager, but MAX reported that the device was not present (Error -88705) and that Windows did not find a driver for the instrument.
We then plugged in a cDAQ-9178 into the fifth cRIO and everything just worked.
What is the difference between a cDAQ-9174 and a cDAQ-9178 other than the number of module supported?
Paul
Solved! Go to Solution.
06-17-2016
06:17 PM
- last edited on
02-10-2025
02:13 PM
by
Content Cleaner
Hi Paul,
In terms of specifications, the two cDAQ chassis should have almost exactly the same specs (at least in all the ways that would affect that connection) -- the only main difference is the number of modules, as you mentioned. More detail on the specs can be found here:
https://www.ni.com/docs/en-US/bundle/cdaq-9174-specs/page/specs.html
https://www.ni.com/en-us/shop/model/cdaq-9178.html
It looks like something may be wrong with your 9174 -- is it working when connected to another system? These are also some great troubleshooting steps to start with: https://knowledge.ni.com/KnowledgeArticleDetails?id=kA00Z000000P7OaSAK&l=en-US
06-19-2016 08:18 AM
Sunayna,
Yes the 9174 works with other computers.
Paul
06-21-2016
04:38 PM
- last edited on
02-10-2025
02:14 PM
by
Content Cleaner
Hi Paul,
This seems like it may be a driver issue! What version of DAQmx do you have installed on both computers? A reinstall of the newest version might be worth a shot since it looks like there may be some corruption. You can find the newest version here: https://www.ni.com/en/support/downloads/drivers/download.ni-daq-mx.html#288248
You can also manually associate the driver with the device, here: https://knowledge.ni.com/KnowledgeArticleDetails?id=kA00Z0000019NFGSA2&l=en-US
Are there any modules plugged into the cDAQ chassis?
Also, sometimes this can happen if you do not reboot your computer after installing NI-DAQmx. Rebooting your computer after installing the driver will allow you to reserve the device successfully. Here are a few other steps to try:
06-22-2016 07:25 AM
Sunayna,
Reloading NI-DAQmx 15.5 and resetting MAX configuration database fixed the issue.
Paul