How to connect dwyer SERIES 629 Differential Pressure Transmitter to the 9219 DAQ and read the signal?

Solved!
Go to solution

Here is my sensor 
Pressure Sensor

Here is the DAQ datasheet:

NI DAQ 9219

Here are my problems/Questions:

First of all I do not know what is LO and HI exactly in the 9219 DAQ device. 

Secondly I do not know which pin I should connect the sensor signal wire to DAQ. Pin 4 or pin 5 ? The sensor has three pins and I assume I should connect the two other wires to the voltage power supply. 
Thirdly how to calibrate the sensor. In labview should I choose voltage in the wizard?

 

I am fairly new in this DAQ and I need your help. 

Thank you

0 Kudos
Message 1 of 2
(3,815 Views)
Solution
Accepted by SilasIII

Hi SilasIII,

 

Hmm well the 3 wires are probably ground, power, and the return signal. The spec sheet for the sensor says this:

Capture.JPG

 

First, you need to know which model you have (4-20mA, 0-5V or 0-10VDC). HI means the return signal, LO means essentialy the ground of the power supply that supplies power to the sensor. Next, you need to get supply for the 13-30 VDC. I don't think that should be too complicated, and can be a simple DC supply from the wall. You might want to learn how to create a custom scale in DAQmx. I hope this provides a starting point.

 

Regards,

Eric