Update -
Although the output of the sensor and the input to the filter/amplifier are _floating_, the output from the filter/amp is [b]single-ended[/b].
I've tried using both DIFF and NRSE with this in mind, but there's very little change in the noise level.
01-22-2009 09:54 AM
Regards,
Dan King
01-22-2009 11:14 AM
Dan - Thanks for the reply!
1) I've tried this in two locations on opposite sides of the building.
2) Yes, there are plenty of fluorescent lights in this room and building. Turned them _all_ off, no change.
3) I will try out aluminum foil today. Do you recommend wrapping everything (BNC cables), or only the unshielded 30AWG leads to the transducer?
4) Unfortunately sampling at 60 Hz is not feasible. Our ultimate goal is a PSD from 5 Hz - 25 kHz, and we'd later like to look at even higher frequencies... 150-200 kHz.
5) True, but this becomes less feasible when I have so many harmonics in the system as well --- check out the example screen capture posted above.
6) Yes, in the differential configuration, I indeed used 100 kOhm resistors between AI+ and GND, and AI- and GND.
If this fully eliminates environment & cabling, I'm beginning to thing that the noise must be generated by the either the Omega power supply or Kemo filter.
01-22-2009 02:17 PM
01-22-2009 04:12 PM
Did you apply the tinfoil with the shiny side IN? 🙂
I jest, these kinds of noise issues can be a bear to deal with. Is there any heavy machinery or electric loads (fans, motors, etc) in the nearby vicinity? AC power conduits/panels in/on the walls?
Do you have access to an oscilloscope? It's best to probe starting out with a mininmal configuration, working from the load cell back up the chain, with and with any DAQ hardware connected.
-AK2DM
01-23-2009 09:42 AM
Regards,
Dan King
01-23-2009 11:43 AM