01-27-2009 08:52 AM
Regards,
Dan King
01-27-2009 09:04 AM
Dan -
Thanks for checking in.
I am in contact with Kemo --- we've arrived at the same conclusion. We're trying to supplement the built-in grounding on the units, and I'm beginning to look for a substitute unit.
- J
01-28-2009 11:04 AM
Regards,
Dan King
06-19-2009 10:39 AM
06-19-2009 01:13 PM
06-19-2009 03:15 PM
I eventually did have a *lot* of luck improving things.
It had a lot to do with improving the quality of my lead wire extensions (shielded twisted pair), and ensuring a common ground- and shield-potential throughout the entire system, including the signal conditioners and DAQ. I used the Kulite's black wire: "excitation -". This was most of my improvement in noise.
In addition, I also built my own signal conditioner (and excitation supply) using ultra-low-noise integrated circuits from TI.
07-14-2009 10:37 AM
Sorry to bother you again. Jerrodh.
Could you please let me know what you did for your own signal conditioner? Add what does the common ground mean? Should I connect AI GND to building ground always? I would really appreciate if you could give me any more sugguestions on this. I'm still have too much noise, like 30 mV.
Did you do something to low your impedance?
Thanks very very much!!!
07-24-2009 01:51 AM
Hi, everyone,
Looks like a lot of people are sufferring.
What I'm facing with is also pretty odd. I could acquire data from one Kulite pressure transducer (XTL-140, 4 wire connection, powered by a 10VDC), but if I use the same power source to power two Kulite transducers (XTL-140 & XTEL-190, connected in parallel), I could get nothing.
Is there anyone ever had the similar problem before?
Thanks,
RD