PCI-6035E registering in MAX as PXI module

Solved!
Go to solution

Hi All,

 

I'm having some trouble properly identifying and using an NI PCI-6035E DAQ. The DAQ is not registering in MAX as a DAQ device but instead as a PXI device. It is listed as PXI4::1::INSTR (PCI-6035E) as shown in the attachment.

 

On the same PC I am running an NI PXI-1033 chassis linked through a PCIe-8361 card. However, even with the PXI system disconnected from the PC the DAQ still registers as a PXI device.

 

Am I running into some sort of conflict between the two devices? We would like to run both together. Currently I'm running MAX 4.3 and LabVIEW 8.5 on Windows7.

 

Any help would be appreciated.

Cheers.

0 Kudos
Message 1 of 3
(2,868 Views)
Solution
Accepted by jnod

How does the PCI-6035E show up in Device Manager? Screenshot of that might be useful if the steps below don't work. 

 

Also, what version of DAQmx do you have? If you don't have DAQmx 9.3, update your device driver by going here and downloading the driver. I'm assuming you are using an operating system that is compatible with that driver. 

 

If you have DAQmx 9.3, try removing the card from your computer, deleting the PXI chassis in MAX and then installing the card again, but to a different PCI slot.

 

Rohama K.
0 Kudos
Message 2 of 3
(2,855 Views)

Installing DAQmx 9.3 seems to have fixed the problem. Thank you.

0 Kudos
Message 3 of 3
(2,847 Views)