Hi,
I am currently working with the PXIe-4467 module for signal generation and have the following questions:
Output Signal Ranges:
The output characteristics of the PXIe-4467 mention signal ranges of ±10 V, ±3.16 V, ±1.0 V, and ±0.316 V.
Sampling Frequency:
If both channels are used simultaneously for signal generating, will this impact the module’s maximum sampling frequency of 250 kS/s?
Looking forward to your clarification. Thank You.
11-20-2024 10:40 PM
You can calculate the resolution/LSB from the bitness and the range, the 4467 uses delta-sigma data converters and it is 24-bit in this case and theoretically, the LSB here for a +/-1V range would be 120nV
Since each channel has a dedicated data converter, you can use the full sampling rate on both channels without any derating.