07-26-2011 11:41 AM
Hi!
I have a NI USB-6212 BNC and it does not show up in the Measurement & Automation Explorer. I am working with LabView 8.0, NI-DAQmx 8.7.1 and Windows XP SP3. In the Device Manager there appears an entry 'USB-621x Loader' when the NI USB-6212 BNC is connected but nothing else happens. What should I do?
Best regards,
Sebastian
08-04-2011 05:49 AM
Hi,
please try to follow this Knowledge base link Why Does My USB DAQ Device Show As USB Firmware Loader in Device Manager?
Hopefully this will help you.
Kind regards
Marek S.
08-04-2011 06:19 AM
Hey Marek,
thanks for your reply. I found out that the USB-6212 BNC differs from the USB-6212. It needs at least DAQmx 8.9. Unfortunately DAQmx 8.9 is not compatible to LabView 8.0.
Best regards,
Sebastian
10-21-2017 07:32 AM
Hi i have same problem. Do u have any suggestions
10-21-2017 12:30 PM
This is a very old thread. Did you read the answers above? Did you try any of the suggestions?
10-22-2017 12:04 AM - edited 10-22-2017 12:06 AM
This is really an old thread - and with a wrong information. You'll need at least DAQmx 9.0 for a USB-6212 BNC. See the readme of that version.
@Rehan23: You really have only access to LabVIEW 8.0??
Regards, Jens