05-03-2016 07:43 AM
I moved a few things around in the rack and was able to see and self test
all devices when connected with my development system but when I connect up
the test PC to this text PXI rack this is what devices can be seen:
What NI Software is missing or what NI software do I need to repair in order to
see the following devices without errors in NI MAX?
6: PXI14::14::INSTR (PXI-4110)
7:PCI Device "PXI114::13::INSTR" (PXI- 4071)
Thank you
05-03-2016 08:24 AM
The PXI-4110 should require the NI-DC Power driver.
The PXI-4071 requires NI-DMM driver.
05-03-2016 09:48 AM
thanks crossrulz-
I am downloading this right now
05-04-2016 07:32 AM
Now if I can only get the teststand sequence files to launch, I can return this PC to our fab house so that
we can start up production again...
05-04-2016 03:21 PM
OK, all devices now appear in NI MAX:
And if I open the test panel on the PXI4071 while running my tesstand sequence file, the front panel
will record a value of 2.45V (which is connected to this device at the time) but the teststand sequence file that is measuring this device reports 0v...
Is there some sort of steps that I need to follow in order to sync the teststand up to the PXI device?
Thanks
05-05-2016 06:41 AM
So it appears that teststand is calling a VI to measure voltages via the PXI-4071 and there are some key files missing like the
DAQmx Write that I need t figure out how to get a copy of that back on this PC...