Hello Nia,
Unfortunately many times error -50400 indicates that the PCI-4461 needs to be repaired. There are a few other things to check before arriving at this conclusion.
1. Has this setup worked in the past or are you just setting this up for the first time?
2. Does the card work when you put it in a different PCI slot in the same computer?
3. Does the card work when you put it in a different computer?
By "work" I mean does the device pass self-test and reset in the Measurement and Automation Explorer (MAX).
If the answer to question one is yes and no for 2 and 3 then you should contact National Instruments directly (using the information at www.ni.com/contact) to obtain a Return Merchandise Authorization. To speed the troubleshooting process when you call in please let the engineer you speak with know about this forum post. They may have a few more questions, but these are also things they'll want to check.
I hope this helps to get you up and running as soon as possible.
Cheers,
04-16-2008 01:37 PM
04-16-2008 05:44 PM
04-17-2008 10:03 AM
Hi NF1
I am going to agree with Brooks on this one. The two errors you are seeing on both computers indicate that the device must be sent in for repair.
Thank you for performing the extra troubleshooting steps, they were very helpful.
Please follow the link that Brooks posted above so you can speak with an engineer and we can start the return process.
Chris_K
09-08-2009 10:22 PM
11-25-2009 10:17 AM
I have a problem with PCI 4461, when i try to reset device show an error
The device selftest has failed. The error report from the device drivers is as follows:
Error -223022 ocurred at self test.
Javier.
11-30-2009 10:20 AM
Ayala,
It is likely that you have a problem with your device. Could you first try these troubleshooting steps before contacting us to repair the device:
03-09-2010 05:29 PM
Hello NF1, Ryan, and Ayala,
Did you all resolve your issues?
We have one PXI-4461, eighty six PXI-4462, all are currently involved in a recall.
Did any of you receive a "Long-Term Fuse Reliability" recall letter from NI?
It mentions errors "-50400, -200313, -200550, and -239022".
Also, in MAX, affected cards, with a blown fuse, will fail the
"Device Reset" from the test panel and report to you error "-223022".
Sounds like you have a blown fuse to me.
Kevin.